Zero problem
Dr. MC Black | 16/03/2021 11:47:34 |
334 forum posts 1 photos | I was given a Digital Micrometer as a gift. It arrived in a Wooden foam lined box. I used it yesterday to measure something having set it to zero by using the ratchet and pressing the reset button After finishing the measurement, I closed the jaws again using the ratchet and the display read 0.002mm That's NOT what I would expect! Am I doing something wrong? Or shall I dispose of the micrometer? I do NOT leave the little button cell installed when the instrument is NOT in use - just in case it leaks! Any suggestions or comments, please? MC Black
|
David George 1 | 16/03/2021 11:59:39 |
![]() 2110 forum posts 565 photos | Hi MC have you thought about the temperatures at this time of year. If you have stored if in a cooler temperature then get it out and use it there could be a difference in readings. The standard temperature is 20 deg C end whilst at work the factory had to kept at this temperature. I would get the micromiter out and give it a while to stabilise to work area temperature before setting up and using it. David |
Journeyman | 16/03/2021 12:05:22 |
![]() 1257 forum posts 264 photos | A 2 micron discrepancy could be caused by temperature (as David says) a minute speck of dust, a slight difference in pressure from the ratchet or almost anything. I am surprised that your micrometer reads down to that level on most the third decimal place is usually only 0 or 5. John |
KWIL | 16/03/2021 12:06:12 |
3681 forum posts 70 photos | This sounds like a digital vernier/caliper, hardly a micrometer. At best these devices are comparators and not for true measurements. |
Howard Lewis | 16/03/2021 12:09:14 |
7227 forum posts 21 photos | When using a digital calliper, I always check the Zero before taking the measurement. 2 microns is not a large error, but you may be finding the limit of resolution of the instrument. It could be the clearance between Thimble and Barrel threads! Fuel injection equipment was made with those sorts of clearances, for 500 - 600 bar pressures, probably even less with current high injection pressures. of 1,500 bar. It could even be inconsistency in the ratchet, or its use. Was the micrometer at the same temperature each time, or was it affected by the heat of your hand, or even your breath? A Standards Room or Calibration Room will not consider measuring anything until it has "soaked" to the ambient temperature of the room for at least 24 hours. Did you turn the ratchet and thimble at the same rate both times? So that the momentum was the same, as far as possible. Taking an extreme example, you won't drive a nail home by pressing the hammer on its head. But you will when it comes flying down at speed! Bear in mind that a human hair is about 125 microns (Varies between 75 and 175 depending on colour ) Howard Edited By Howard Lewis on 16/03/2021 12:11:29 |
Bo'sun | 16/03/2021 12:12:59 |
754 forum posts 2 photos | Good afternoon MC, As David says. Temperature could well be the culprit. After all, 0.002mm (0.00008" I think I'd be happy with that. Certainly for model engineering. |
Michael Gilligan | 16/03/2021 12:23:09 |
![]() 23121 forum posts 1360 photos | Posted by Howard Lewis on 16/03/2021 12:09:14:
. 2 microns is not a large error […] Bear in mind that a human hair is ...
. From a previous discussion:
Here’s a useful ‘benchmark’ image, courtesy of Wikipedia: .... a 50 micron human hair, and a 6 micron carbon filament in the same frame: https://en.wikipedia.org/wiki/Micrometre#/media/File:Cfaser_haarrp.jpg . MichaelG. |
Dr. MC Black | 16/03/2021 12:26:37 |
334 forum posts 1 photos | Posted by David George 1 on 16/03/2021 11:59:39:
Hi MC have you thought about the temperatures at this time of year. If you have stored if in a cooler temperature then get it out and use it there could be a difference in readings. The standard temperature is 20 deg C end whilst at work the factory had to kept at this temperature. I would get the micromiter out and give it a while to stabilise to work area temperature before setting up and using it. David Thank you for taking the time to respond. The Micrometer was stored indoors (in my kitchen) where I did the measuring. So I don't think the difference was due to temperature - but it's a good thought. MC
|
Tony Pratt 1 | 16/03/2021 12:35:45 |
2319 forum posts 13 photos | 0.002 mm is nothing to worry about. Tony |
Oily Rag | 16/03/2021 12:37:08 |
![]() 550 forum posts 190 photos | I was taught during my apprenticeship never to use the ratchet on a micrometer, always to 'feel' the contact point at which to take the reading. The reason? ratchets are inconsistent and not a consistent way to measure anything, better to develop the 'feel'. |
Mike Poole | 16/03/2021 13:16:03 |
![]() 3676 forum posts 82 photos | Did you clean the micrometer faces before zeroing? I use a piece of clean paper gripped lightly between the faces and then drawn gently out of the faces. Be sure to not grip too tightly and produce paper dust. With the anvil faces clean, close and zero. If you can indicate thousandths of a millimetre then your technique is going to be very good to get repeatable readings. If you have access to slip gauges then measure some and see if your technique with the ratchet or thimble agrees with the slip gauge, you will be doing well if you get repeatable readings and if the reading agrees perfectly with the slips then you have a fantastic mic. Mike Edited By Mike Poole on 16/03/2021 13:16:46 Edited By Mike Poole on 16/03/2021 13:18:13 |
Dr. MC Black | 16/03/2021 13:52:27 |
334 forum posts 1 photos | Posted by Oily Rag on 16/03/2021 12:37:08:
I was taught during my apprenticeship never to use the ratchet on a micrometer, always to 'feel' the contact point at which to take the reading. The reason? ratchets are inconsistent and not a consistent way to measure anything, better to develop the 'feel'. Thank you for taking the time to respond. I was taught, over 50 years ago, to always use the ratchet! I don't use a micrometer sufficiently often to develop a feel!
MC |
Dr. MC Black | 16/03/2021 14:07:34 |
334 forum posts 1 photos | Posted by Mike Poole on 16/03/2021 13:16:03:
Did you clean the micrometer faces before zeroing? I use a piece of clean paper gripped lightly between the faces and then drawn gently out of the faces. Be sure to not grip too tightly and produce paper dust. With the anvil faces clean, close and zero. If you can indicate thousandths of a millimetre then your technique is going to be very good to get repeatable readings. If you have access to slip gauges then measure some and see if your technique with the ratchet or thimble agrees with the slip gauge, you will be doing well if you get repeatable readings and if the reading agrees perfectly with the slips then you have a fantastic mic. Mike Thank you for taking the time to respond. I didn't think to clean the faces. I'll try to remember about doing that in the future but would expect the original reading to be greater than the final reading if the faces had been dirty. The only calibration opportunities that I have are one inch and 3 inch standards - neither much help with a 0-25mm instrument! MC
|
Emgee | 16/03/2021 14:08:26 |
2610 forum posts 312 photos | I follow the same cleaning method as Mike and check occasionally on a slip, but to get used to using the mic with the same force you can use an HSS or carbide tool shank, when you get the same reading 8 out of 10 you will be confident measuring parts. Emgee |
Michael Gilligan | 16/03/2021 14:19:18 |
![]() 23121 forum posts 1360 photos | Posted by MC Black on 16/03/2021 14:07:34: […] I didn't think to clean the faces. I'll try to remember about doing that in the future but would expect the original reading to be greater than the final reading if the faces had been dirty. . . Unfortunately, it is entirely plausible that a two micron dust particle landed on one anvil during the time you were conducting the test ... unless, of course, you have full HEPA filtration throughout your working environment. MichaelG. . Edit, for context: https://www.ristenbatt.com/xcart/Vacuum-Cleaner-Filtration-Efficiency.html Edited By Michael Gilligan on 16/03/2021 14:20:31 |
not done it yet | 16/03/2021 14:21:03 |
7517 forum posts 20 photos | Posted by Journeyman on 16/03/2021 12:05:22:
A 2 micron discrepancy could be caused by temperature (as David says) a minute speck of dust, a slight difference in pressure from the ratchet or almost anything. I am surprised that your micrometer reads down to that level on most the third decimal place is usually only 0 or 5. John Agreed, John, re the temperature - even in a kitchen environment. Mine resolves to 0.001mm - but who on here really needs that third figure to be absolutely accurate? (Only Howard with fuel injection equipment, maybe?) I think mine was likely rejected due to a tiny chip on one anvil edge. I find using the ratchet is close enough for me! IF that micrometer is only a 0-25mm, the error is not great. If it happens to be a 25-50mm, or larger, that error becomes virtually non-existent - even if there is a similar error at the reading value! I don’t think I would be giving MC a horse as a gift.🐎 |
Dr. MC Black | 16/03/2021 16:10:02 |
334 forum posts 1 photos | Posted by not done it yet on 16/03/2021 14:21:03:
I don’t think I would be giving MC a horse as a gift.🐎
Thank you for taking the time to respond. I don't want a horse, Thank you. They are nasty smelly creatures that bite at one end and fertilise rhubarb at the other. I'm NOT sure that my steel toecap safety boots could withstand even a stationary horse on them. With best wishes and thanks again. MC
|
Peter Greene | 16/03/2021 17:29:06 |
865 forum posts 12 photos | Posted by MC Black on 16/03/2021 11:47:34:
After finishing the measurement, I closed the jaws again using the ratchet and the display read 0.002mm That's NOT what I would expect!
oooh I wish my digital mic could measure to that level of accuracy and repeatability. I'm sure I could dream up a use for it if it could. .... oh .... wait ..... we're not talking accuracy or repeatability we're talking digital display reading. Ah well ... |
Dave Halford | 16/03/2021 19:00:00 |
2536 forum posts 24 photos | What do you make MC? |
Dr. MC Black | 17/03/2021 16:50:19 |
334 forum posts 1 photos | I have inherited a lot of tools et hoc genus omne from my late father (and another friend) and have been spending warmer days during the Covid-19 pandemic lockdown restoring them When my workshop is sorted out, I look forward to making some model engines I recall seeing in an Axminster catalogue several years ago a kit for an engine suitable for beginners but it's NOT in their current catalogue Does anybody have the drawings for that which they would be prepared to share, please? (Perhaps I should have started a new thread for that?) With best wishes. MC
|
Please login to post a reply.
Want the latest issue of Model Engineer or Model Engineers' Workshop? Use our magazine locator links to find your nearest stockist!
Sign up to our newsletter and get a free digital issue.
You can unsubscribe at anytime. View our privacy policy at www.mortons.co.uk/privacy
You can contact us by phone, mail or email about the magazines including becoming a contributor, submitting reader's letters or making queries about articles. You can also get in touch about this website, advertising or other general issues.
Click THIS LINK for full contact details.
For subscription issues please see THIS LINK.